TAYLOR SPATIAL FRAME
Hexapod System | Stewart Platform | 6-Axis Correction
CORRECTION PARAMETERS
Critical Must-Knows
- Hexapod = Stewart platform: 6 telescopic struts connecting 2 rings = 6 DOF
- Reference ring: Fixed ring (usually proximal) defines coordinate system
- Moving ring: Ring that moves relative to reference ring during correction
- Mounting parameters: 14 measurements (7 per ring) required for software
- Total residual: Combined measure of all deformity components - correction complete when approaches zero
Examiner's Pearls
- "Stewart platform originally designed for flight simulators (1965)
- "TSF can correct all 6 axes simultaneously - unlike Ilizarov
- "Virtual hinge at CORA minimizes unwanted translation during angular correction
- "Frame mounting errors cause residual deformity - precision critical
Clinical Imaging
Imaging Gallery

Critical TSF Exam Points
Hexapod Principle
Stewart platform kinematic mechanism: 6 telescopic struts connect 2 rings with universal joints. This configuration provides 6 degrees of freedom (3 translations + 3 rotations), allowing simultaneous multiplanar correction. Unlike Ilizarov, no hinge placement or frame rebuilding required.
Mounting Parameters
14 total parameters (7 per ring) must be measured accurately: ring diameter, ring-to-ring distance, frame offset (anterior, lateral), strut mounting positions (rotational orientation), and axial orientation. Errors in measurement cause residual deformity.
Reference vs Moving Ring
Reference ring = fixed ring (typically proximal) that defines the coordinate system. Moving ring = ring that moves during correction. Choice affects deformity prescription - be consistent. The reference ring is attached to the "stable" bone segment.
Total Residual
Total residual = composite measure of remaining deformity across all 6 axes. Calculated by software after inputting current deformity parameters. Correction progresses until total residual approaches zero. Used to monitor progress and endpoint.
Taylor Spatial Frame vs Ilizarov Circular Fixator
| Feature | Taylor Spatial Frame (TSF) | Ilizarov Circular Fixator |
|---|---|---|
| Strut configuration | 6 telescopic struts (hexapod) | Rods, hinges, motors (modular) |
| Degrees of freedom | 6 DOF simultaneous correction | Sequential corrections required |
| Planning method | Computer software (web-based) | Mechanical/manual planning |
| Hinge placement | Virtual hinge (software) | Physical hinges required |
| Frame rebuilding | Not required during treatment | Often needed for direction changes |
| Learning curve | Software-dependent (steep initially) | Traditional mechanics (steep) |
| Cost | Higher (struts, software) | Lower (traditional components) |
| Residual correction | Easy re-prescription via software | Frame reconfiguration needed |
TAR-VAR6 Degrees of Freedom
Memory Hook:TAR-VAR = 6 DOF: 3 Translations (AP, Axial, ML) + 3 Rotations (Varus/valgus, Axial rotation, Recurvatum)!
DR FLOSSMounting Parameters (7 per ring)
Memory Hook:DR FLOSS your frame: Diameter, Ring distance, Frame offset (AP), Lateral offset, Orientation, Strut positions, Segment!
MAD CORADeformity Parameters (CORA Analysis)
Memory Hook:MAD CORA analysis: Mechanical axis, Angulation, Displacement + CORA, Orientation, Rotation, Axial length!
Overview and Principles
The Taylor Spatial Frame (TSF) is a hexapod external fixator based on the Stewart platform mechanism, originally developed by Dr. J. Charles Taylor in Memphis, Tennessee. It enables simultaneous six-axis deformity correction through computer-assisted planning and graduated strut adjustments.
Key Principles:
The TSF consists of two rings connected by six telescopic struts arranged in a specific geometric pattern. This hexapod configuration creates a Stewart-Gough platform - the same kinematic mechanism used in flight simulators, robotic surgery, and precision positioning systems.
Six Degrees of Freedom:
- Three translational: Anterior-posterior, medial-lateral, axial (shortening/lengthening)
- Three rotational: Varus/valgus (coronal angulation), flexion/extension (sagittal angulation), axial rotation (torsion)
Advantages over Traditional Ilizarov:
- Simultaneous multiplanar correction - no need for sequential corrections
- No physical hinges required - virtual hinge calculated by software
- No frame rebuilding - same construct throughout treatment
- Residual correction capability - easily re-prescribe if deformity persists
- Computer-assisted planning - reduces human calculation errors
Stewart Platform History
The Stewart platform was described by D. Stewart in 1965 for use in flight simulation. It provides 6 degrees of freedom through 6 linear actuators connecting two platforms. Taylor applied this mechanism to orthopaedic external fixation in the 1990s, revolutionizing deformity correction.
Indications and Contraindications
Indications
Deformity Correction:
- Complex multiplanar deformities (congenital, developmental, post-traumatic)
- Tibial and femoral malunion
- Metabolic bone disease deformities (rickets, Blount disease)
- Angular deformities with rotational and translational components
Limb Lengthening:
- Limb length discrepancy requiring distraction osteogenesis
- Combined lengthening with deformity correction
- Congenital short femur, fibular hemimelia
Fracture Management:
- Acute fracture reduction with gradual correction
- Malunion correction with osteotomy
- Nonunion with deformity
Infection and Bone Loss:
- Bone transport for segmental defects
- Infected nonunion management
- Combined with Masquelet technique
Contraindications
Absolute:
- Active sepsis (systemic)
- Severe soft tissue compromise precluding pin/wire placement
- Non-compliant patient (unable to perform daily adjustments)
- Inadequate bone stock for wire/pin fixation
Relative:
- Severe peripheral vascular disease
- Uncontrolled diabetes mellitus
- Osteoporosis (may require modified technique)
- Morbid obesity (frame stability concerns)
- Psychological unsuitability
Technical Principles
The Hexapod Mechanism
Stewart Platform Kinematics:
The TSF utilizes inverse kinematics - given a desired end position/orientation, the software calculates the required strut lengths. This is the opposite of forward kinematics (calculating position from joint angles).
Key Components:
- Two rings: Reference ring and moving ring
- Six struts: Telescopic struts with universal joints at each end
- Universal joints: Allow multi-axis rotation at strut-ring connections
- Strut mechanism: Threaded telescoping design for length adjustment
Why 6 Struts?
Six struts provide exactly 6 degrees of freedom - the minimum required for full spatial positioning. This is mathematically analogous to 6 linear equations with 6 unknowns, creating a fully determined (not over- or under-constrained) system.
Reference and Moving Rings
Reference Ring:
- The "fixed" ring that defines the coordinate system
- Usually placed on the proximal (stable) segment
- All deformity parameters measured relative to this ring
- Convention: Typically the ring closest to the trunk
Moving Ring:
- The ring that moves during correction
- Attached to the distal (deformed) segment
- Moves to the corrected position as struts adjust
Reference Ring Selection
The choice of reference ring affects how deformity is prescribed to the software. If you select the proximal ring as reference, describe the deformity of the distal segment. Be consistent - errors in reference ring selection lead to correction in the wrong direction.
Virtual Hinge Concept
In traditional Ilizarov correction, physical hinges are placed at the Center of Rotation of Angulation (CORA) to achieve pure angular correction without translation.
The TSF creates a virtual hinge through software calculation:
- The software calculates strut adjustments that simulate rotation about a hinge at any specified point
- No physical hinge needed
- Virtual hinge placed at CORA minimizes secondary translation
- Can be adjusted/repositioned during treatment if needed
CORA (Center of Rotation of Angulation):
- The intersection of the proximal and distal mechanical axes
- Ideal pivot point for angular correction
- Correction about CORA produces pure angulation without translation
- Correction about a point away from CORA produces combined angulation and translation
Total Residual
Definition:
The total residual is a composite measure representing the magnitude of remaining deformity across all 6 axes after software calculation.
Calculation:
- Software combines angular deformities (degrees) and translational deformities (mm)
- Weighted combination into single value
- Approaches zero as correction completes
Clinical Use:
- Monitor correction progress
- Endpoint determination (typically less than 5mm equivalent)
- Identify incomplete corrections requiring adjustment
Surgical Technique
Preoperative Planning
Imaging:
- Long-leg standing radiographs (AP and lateral)
- CT scan for rotational deformity assessment
- EOS imaging if available (low radiation, full-length)
Deformity Analysis:
- Mechanical axis deviation (MAD)
- Identify CORA(s) - may be single or multiple
- Measure angular deformity (magnitude and direction)
- Assess translational deformity
- Evaluate rotational deformity (CT required)
- Measure limb length discrepancy
Osteotomy Planning:
- Level of osteotomy (ideally at CORA)
- If CORA outside bone, plan for secondary correction
- Corticotomy technique for lengthening
Ring Selection and Sizing
Ring Diameter:
- Allow 2-3 finger breadths clearance circumferentially
- Consider soft tissue swelling
- Larger rings allow better access but reduce stability
Ring Configuration:
- Full rings vs 5/8 rings vs foot plates
- Proximal tibial: Consider knee ROM
- Distal tibial: Consider ankle clearance
- Femoral: May need half-rings for sitting
Frame Assembly
Steps:
- Assemble rings with appropriate connectors
- Attach 6 struts to designated mounting positions
- Ensure universal joints move freely
- Check strut length range adequate for planned correction
- Record strut positions on mounting worksheet
Wire and Pin Placement
General Principles:
- Minimum 2 fixation elements per ring (wires and/or pins)
- Tensioned wires (1.5-1.8mm) for ring stability
- Half-pins (5-6mm) for added stability
- Safe corridors to avoid neurovascular structures
Tibial Application:
- Proximal ring: Safe wire from fibular head to posteromedial tibia
- Reference wire perpendicular to tibial axis
- Avoid peroneal nerve
- Distal ring: Safe zone anterior to posterior
- Protect anterior tibial vessels
Wire Tensioning:
- Olive wires: 130-150 kg tension
- Smooth wires: 110-130 kg tension
- Re-tension after 48 hours (settling)
This section covers frame application technique.
Complications
Frame-Related Complications
Pin Site Infection:
- Most common complication (30-100% incidence)
- Superficial: Local care, oral antibiotics
- Deep: May require pin removal, IV antibiotics
- Prevention: Regular pin care, good technique
Pin Loosening:
- Presents as pain, instability, increased discharge
- May require pin replacement
- More common with half-pins in poor bone
Frame Instability:
- Due to pin failure, ring breakage, strut malfunction
- Requires urgent assessment and revision
Correction-Related Complications
Residual Deformity:
- Due to mounting parameter errors
- Incorrect deformity prescription
- Patient non-compliance with schedule
- Management: Re-measure, re-prescribe
Over-correction:
- Deformity corrected beyond neutral
- May be intentional (planned over-correction for rebound)
- Unintentional: Revise prescription
Neurovascular Injury:
- Stretch neuropathy with lengthening
- Vascular compromise (rare)
- Slow or stop distraction if symptoms develop
Bone-Related Complications
Premature Consolidation:
- Regenerate consolidates before target length
- Management: Speed up distraction, consider re-osteotomy
Delayed Consolidation/Nonunion:
- At osteotomy site or regenerate
- Risk factors: Smoking, diabetes, poor technique
- Management: Bone graft, optimize biology
Regenerate Fracture:
- After frame removal
- Protect with cast/brace until remodeling complete
Soft Tissue Complications
Joint Stiffness:
- Adjacent joints stiffen during prolonged treatment
- Prevention: Aggressive physiotherapy
- May require manipulation, release
Muscle Contracture:
- Gastrocnemius (equinus) in tibial lengthening
- Quadriceps in femoral lengthening
- Consider prophylactic releases for large lengthenings
Complication Prevention
The most significant functional complications are joint stiffness and contracture. Physiotherapy from day one is essential. Monitor ROM at every visit. Consider prophylactic soft tissue releases (gastrocnemius slide, quadricepsplasty) for lengthenings greater than 4-5cm.
Evidence Base
Taylor Spatial Frame vs Ilizarov for Tibial Deformity
- TSF achieved equivalent correction accuracy to Ilizarov (mean 98% correction)
- TSF required significantly fewer frame adjustments during treatment
- No difference in union rates or time to frame removal
- TSF learning curve offset by software assistance
Mounting Parameter Accuracy in TSF
- Frame offset measurements most prone to error
- 5mm error in offset can produce 5 degrees residual angulation
- Standardized measurement protocol reduces errors
- Software residual mode allows correction of errors mid-treatment
Computer-Assisted Deformity Correction Outcomes
- 95% achieved planned correction
- Mean correction time 3.2 months for angular deformity
- Pin site infection rate 32% (all superficial)
- No major neurovascular complications
Hexapod Frame for Fracture Reduction
- Gradual reduction achieved in all cases
- Avoids risks of acute open reduction
- Particularly useful for periarticular fractures with soft tissue compromise
- Union achieved in 95% without bone grafting
Exam Viva Scenarios
Practice these scenarios to excel in your viva examination
Scenario 1: TSF Principles
"Explain the Taylor Spatial Frame to me. How does it work and what advantages does it offer over traditional Ilizarov frames?"
Scenario 2: Complex Tibial Deformity
"A 25-year-old man presents with a tibial malunion following conservative treatment of a fracture 2 years ago. Radiographs show 20 degrees varus angulation, 15 degrees procurvatum, and 2cm shortening. How would you plan correction using a TSF?"
Scenario 3: Residual Deformity After TSF
"You applied a TSF for correction of a tibial deformity. At the end of the planned correction schedule, radiographs show a 10-degree residual varus deformity. What went wrong and how would you manage this?"
Australian Context
In Australia, the Taylor Spatial Frame is utilized at specialized limb reconstruction units within major metropolitan hospitals. The technology represents an important advance in managing complex limb deformities and is increasingly preferred for multiplanar corrections.
Device Availability and Training: The TSF is available through Smith and Nephew, with dedicated software platforms for case planning. Australian orthopaedic surgeons undertaking limb reconstruction fellowships receive training in TSF application, and the AO Foundation and ASAMI (Association for the Study and Application of Methods of Ilizarov) provide courses on hexapod frame application. The learning curve is significant, but computer-assisted planning reduces technical errors compared to traditional Ilizarov techniques.
Clinical Applications: Australian centres use the TSF for complex tibial and femoral deformity correction, limb lengthening, and fracture management. The ability to correct multiplanar deformities simultaneously is particularly valuable in post-traumatic malunion cases and congenital limb deficiencies. Combined with distraction osteogenesis principles, the TSF is used for limb length discrepancy management in conditions such as fibular hemimelia and congenital short femur.
Practical Considerations: Patient selection is critical, as TSF treatment requires significant compliance for daily strut adjustments and pin site care over prolonged periods (typically 3-6 months in frame). Multidisciplinary input including limb reconstruction surgeons, specialized physiotherapists, and orthotists is essential. The cost of the TSF system is higher than traditional Ilizarov components, but this is offset by reduced need for frame modifications and potentially shorter learning curve for multiplanar corrections.
TAYLOR SPATIAL FRAME
High-Yield Exam Summary
Hexapod Principles
- •Stewart platform: 6 struts, 2 rings, 6 DOF
- •3 translational: AP, ML, axial
- •3 rotational: varus/valgus, flexion/extension, rotation
- •Virtual hinge at CORA (no physical hinge)
Mounting Parameters (14 total)
- •7 per ring: diameter, ring distance, AP offset, lateral offset
- •Master tab position, strut positions, segment/side
- •Errors cause residual deformity
- •Standardized measurement protocol essential
Reference vs Moving Ring
- •Reference ring: stable segment (usually proximal)
- •Moving ring: deformed segment (moves during correction)
- •Defines coordinate system for software
- •Be consistent with ring selection
Software Planning
- •Input 14 mounting parameters
- •Prescribe deformity in all 6 axes
- •Software calculates daily strut changes
- •Residual mode for mid-treatment adjustment
TSF vs Ilizarov Advantages
- •Simultaneous multiplanar correction
- •No hinge placement or frame rebuilding
- •Computer-assisted (reduces errors)
- •Easy residual correction via software
Distraction Protocol
- •Latency: 5-7 days before starting
- •Rate: 1mm/day in 4 divided doses
- •Angular correction: 1-2 degrees/day
- •Total residual approaches zero when complete