Fluoroscopy Principles

Back to Topics
Contents
0%

Fluoroscopy Principles

Comprehensive guide to fluoroscopy principles covering image generation, radiation exposure, dose reduction strategies, and intraoperative orthopaedic applications for fellowship exam preparation.

Very High Yield
complete
Reviewed: 2026-03-11By OrthoVellum Medical Education Team

Reviewed by OrthoVellum Editorial Team

Orthopaedic clinicians and medical editors β€’ Published by OrthoVellum Medical Education Team

Editorial boardMethodologyReview policyReport a correction
High Yield Overview

Fluoroscopy Principles

Real-Time Intraoperative Imaging for Orthopaedic Surgery

C-armStandard intraoperative device
30fpsMaximum frame rate (continuous)
1-5 mSvTypical procedure dose to patient
PulsedMode reduces dose by up to 90%
II/FPDImage intensifier / flat panel detector
ALARACore radiation safety principle
InverseΒ²Dose decreases with distance squared
LeadMinimum 0.5mm Pb equivalence for aprons

Fluoroscopy Modes

Continuous: Constant X-ray output β€” highest dose, smoothest image (rarely needed)

Pulsed: Intermittent X-ray pulses (1-30 per second) β€” significant dose reduction, adequate image quality

Last Image Hold: Stores the last frame for review β€” zero additional radiation

Screening (low-dose): Reduced tube output for positioning β€” image quality reduced but acceptably so

Cine acquisition: High-dose recording for documentation β€” minimal use in orthopaedics

Key: Pulsed fluoroscopy at low pulse rates with last image hold should be the default mode for all orthopaedic procedures

Critical Must-Knows

  • Fluoroscopy is continuous or pulsed X-ray imaging providing real-time visualisation during procedures.
  • The C-arm consists of an X-ray tube (below the patient) and an image receptor (above) connected by a C-shaped frame.
  • Scatter radiation to the surgeon is greatest on the X-ray tube side β€” the surgeon should stand on the IMAGE RECEPTOR side.
  • Pulsed fluoroscopy reduces radiation dose by up to 90% compared to continuous fluoroscopy.
  • The inverse square law: doubling the distance from the X-ray source reduces dose to one-quarter.

Examiner's Pearls

  • "
    The X-ray tube is the source of SCATTER radiation β€” always position it AWAY from the surgeon (usually under the table).
  • "
    Last image hold (LIH) stores the last fluoroscopic frame, reducing the need for additional exposure.
  • "
    Mini C-arm (extremity fluoroscopy) produces significantly lower radiation than a standard C-arm.
  • "
    Magnification mode increases dose because the X-ray beam is collimated more tightly and tube output increases.
  • "
    Lead thyroid shield reduces thyroid dose by 90% and is mandatory for all fluoroscopy-exposed personnel.

Exam Warning

Fluoroscopy principles are examined extensively in both physics viva stations and clinical/operative scenarios. You must understand: the components and operation of a C-arm, scatter radiation and positioning (tube side vs detector side), dose reduction strategies (ALARA, pulsed mode, collimation, LIH), the inverse square law, magnification effect on dose, and the occupational dose limits for radiation workers. A classic trap is placing the surgeon on the X-ray tube side.

Mnemonic

CLAPTDRadiation Dose Reduction

C
Collimation
Narrow the beam to the region of interest β€” reduces irradiated volume and scatter production
L
Last image hold
Use stored images for review instead of firing additional X-ray exposures
A
ALARA and As low as possible kV/mA
As Low As Reasonably Achievable β€” use the minimum technique factors that produce adequate image quality
P
Pulsed fluoroscopy
Use pulsed mode at the lowest acceptable pulse rate (1-4 pulses/second reduces dose by 50-90%)
T
Time minimisation
Use fluoroscopy only when needed. Release the pedal when not actively screening. Brief exposures rather than continuous
D
Distance maximisation
Inverse square law: doubling distance from the source reduces dose to one-quarter. Stand as far from the X-ray tube as practical

Memory Hook:CLAPTD to reduce dose: Collimate, Last image hold, ALARA, Pulsed mode, Time reduction, Distance.

Mnemonic

SAFEC-arm Positioning

S
Source (tube) UNDER the table
Standard orientation: X-ray tube below, image receptor above. This directs the primary beam upward and scatter downward
A
Away from the tube = less scatter to surgeon
The surgeon should stand on the IMAGE RECEPTOR (detector) side, away from the X-ray tube which is the major scatter source
F
Face the receptor
The surgeon's body faces the image receptor side β€” the side with LESS scatter radiation
E
Entry beam at tabletop = most scatter
The X-ray beam entering the patient produces the most scatter at the entry surface β€” keep this below the table

Memory Hook:SAFE positioning: Source Under, Away from tube, Face receptor, Entry below β€” protect yourself from scatter.

Mnemonic

MAGWhen Fluoroscopy Dose Increases

M
Magnification mode
Magnification collimates the beam more tightly and the system compensates by increasing tube output β€” dose increases 2-4x
A
Automatic brightness control (ABC)
ABC automatically increases X-ray output as patient thickness increases β€” obese patients receive significantly more dose
G
Greater patient thickness
Larger patients require more radiation to penetrate the tissues, and more scatter is produced β€” dose to patient and staff both increase

Memory Hook:MAG causes dose to go up: Magnification, Automatic brightness control, and Greater patient thickness.

Overview

Fluoroscopy is the most commonly used intraoperative imaging modality in orthopaedic surgery. It provides real-time X-ray imaging that enables accurate fracture reduction, implant placement, joint assessment, and screw trajectory verification during surgical procedures. The mobile C-arm is a fundamental tool in every orthopaedic operating theatre.

Understanding the physics of fluoroscopy and the principles of radiation dose reduction is essential for safe practice. Orthopaedic surgeons are among the highest-exposed medical professionals to occupational radiation, and the cumulative effect of years of fluoroscopy exposure must be actively minimised through proper technique and protective equipment.

Key Components of a C-arm

The C-arm consists of: (1) X-ray tube β€” generates the X-ray beam (cathode/anode system). (2) Image receptor β€” either an image intensifier (older) or flat-panel detector (newer). (3) C-shaped gantry β€” connects tube and receptor, allowing rotation in multiple planes. (4) Monitor β€” displays the real-time image. (5) Pedal controls β€” foot-operated for start/stop of exposure. (6) Collimator β€” adjustable shutters that narrow the beam. A flat-panel detector (FPD) C-arm provides better image quality, less dose, and less distortion compared to an image intensifier (II) C-arm.

ALARA Principle

ALARA (As Low As Reasonably Achievable) is the foundational principle of radiation protection. Every exposure should use the minimum dose necessary to achieve adequate image quality for the clinical task. This applies to both the patient and the operating team. Practical implementation: pulsed mode, collimation, distance, shielding, last image hold, minimising screening time.

Clinical Imaging

Imaging Gallery

C-arm fluoroscopy unit demonstrating standard intraoperative positioning for orthopaedic surgery
Click to expand
A C-arm fluoroscopy unit in the standard intraoperative configuration. The X-ray tube is positioned below the patient (reducing scatter to the operating team), with the image receptor above. The C-shaped gantry allows rotation to obtain AP, lateral, and oblique views without repositioning the patient.Credit: Open-i (NIH) (Open Access (CC BY))
Fluoroscopic image demonstrating intraoperative orthopaedic hardware assessment
Click to expand
Intraoperative fluoroscopic image demonstrating real-time assessment of orthopaedic hardware positioning. The ability to verify fracture reduction, implant position, and screw trajectory in real time is the fundamental advantage of fluoroscopy over other imaging modalities during surgery.Credit: Open-i (NIH) (Open Access (CC BY))

Systematic Approach

Systematic Dose Reduction in the Operating Theatre

Intraoperative Dose Reduction Framework

StrategyImplementationExpected Dose Reduction
Pulsed fluoroscopyUse lowest acceptable pulse rate (1-4 pulses/second for positioning; 8-15 for dynamic assessment)50-90% reduction compared to continuous mode
Last image holdReview stored images instead of firing additional exposuresEliminates unnecessary repeat exposures β€” cumulative benefit
CollimationNarrow the beam to the exact region of interest using the shuttersReduces irradiated volume AND scatter by 30-50%
Distance maximisationStep back from the beam when not actively operating; extend hands from beam pathDoubling distance = 75% dose reduction (inverse square law)
Minimise screening timeUse brief exposures (less than 2 sec each); release pedal between checksDirect proportional reduction β€” halving time halves dose
Avoid magnificationUse magnification mode only when absolutely necessary for detailEliminates 2-4x dose increase associated with magnification

Critical Exam Pearl: Scatter Radiation

Scatter radiation is the PRIMARY source of occupational dose to the operating team. The X-ray tube is the main source of scatter β€” scatter is produced when the primary beam enters the patient. Therefore: (1) The surgeon should ALWAYS stand on the IMAGE RECEPTOR (detector) side, not the X-ray tube side. (2) With the standard C-arm orientation (tube below, receptor above), the surgeon stands on the side AWAY from under the table. (3) Scatter intensity decreases rapidly with distance from the patient (approximately inversely with the square of the distance). (4) Lead aprons attenuate approximately 90-95% of scatter radiation at orthopaedic energies.

Physics and Equipment

X-ray Production in Fluoroscopy

The X-ray tube contains a cathode (tungsten filament) heated by electrical current, which produces electrons by thermionic emission. These electrons are accelerated by a high voltage (kV) across a vacuum toward the anode (tungsten target). When the electrons strike the anode, approximately 99% of their kinetic energy is converted to heat and only 1% produces X-rays.

Two processes produce X-rays:

  • Bremsstrahlung (braking) radiation: Electrons decelerated by the nuclear electric field of tungsten atoms produce a continuous spectrum of X-ray energies. This is the dominant process.
  • Characteristic radiation: Electrons knock out inner-shell electrons from tungsten atoms; the resulting cascade produces X-rays at specific energies (characteristic of tungsten).

Key tube parameters:

  • kV (kilovoltage): Controls the ENERGY (penetrating power) of the X-ray beam. Higher kV = more penetrating beam. Typical fluoroscopy: 60-120kV.
  • mA (milliamperage): Controls the NUMBER of X-rays produced per unit time. Higher mA = more X-rays = brighter image but more dose.
  • Automatic Brightness Control (ABC): The system automatically adjusts kV and mA to maintain consistent image brightness as patient thickness varies. This is why obese patients receive significantly more radiation β€” the system increases output to compensate for greater tissue attenuation.

Image Intensifier vs Flat-Panel Detector

Image Intensifier (II): The older technology. The X-ray beam passes through the patient and strikes an input phosphor (caesium iodide) on a curved vacuum tube, producing light. This light generates electrons at a photocathode, which are accelerated and focused onto an output phosphor, producing a visible image. Key characteristics: causes pincushion distortion (peripheral image warping), S-distortion from earth's magnetic field, vignetting (reduced brightness at edges), and veiling glare (reduced contrast due to internal scatter).

Flat-Panel Detector (FPD): Modern technology. A matrix of amorphous silicon photodiode detectors directly converts X-rays to digital signal. Key advantages over II: NO pincushion distortion, better dynamic range, better contrast resolution, faster readout, lower dose for equivalent image quality, and more compact design. FPDs are standard on modern C-arms and have largely replaced image intensifiers.

Mini C-arm: A smaller, dedicated extremity fluoroscopy unit operating at lower kV and mA. Produces significantly less radiation than a standard C-arm (approximately one-tenth the dose for hand and wrist procedures). Ideal for distal extremity procedures where its smaller field of view is sufficient.

Intraoperative Applications

Common Fluoroscopy Applications in Orthopaedic Surgery

ApplicationTypical ViewsKey Technical Points
Fracture fixation (long bones)AP and lateral of fracture site; inlet/outlet for pelvisVerify reduction, alignment, and implant position. Use traction views for length assessment
Intramedullary nailingAP and lateral at fracture and at proximal/distal locking sitesFreehand interlocking screw technique relies on 'perfect circle' fluoroscopic alignment of the screw hole
Hip fracture fixation (DHS/IM nail)AP and lateral hip viewsEnsure tip-apex distance (TAD) is less than 25mm on combined AP and lateral views
Pedicle screw insertionAP, lateral, and oblique views of the spineAP view confirms medial/lateral position; lateral confirms depth and angulation
Joint replacementAP and lateral views of the jointVerify component position, alignment, and cement distribution in cemented arthroplasty
K-wire and pin placementAP and lateral of the target areaReal-time guidance for percutaneous fixation. Use mini C-arm for distal extremity

The Perfect Circle Technique

For freehand interlocking screw placement in intramedullary nailing, the C-arm is rotated until the nail hole appears as a 'perfect circle' on the fluoroscopic image. This means the X-ray beam is perfectly aligned with the axis of the hole. A drill or screw inserted toward the centre of this circle will be perfectly aligned with the hole axis. This technique eliminates the need for jig-based targeting systems and reduces radiation exposure because it minimises the number of screening attempts needed.

Evidence Base

Radiation Dose to Orthopaedic Surgeons During Fluoroscopy

Systematic Review
Giordano BD, Baumhauer JF, Morgan TL, Rechtine GR β€’ Journal of Bone and Joint Surgery (American) (2011)
Key Findings:
  • Orthopaedic surgeons receive a mean effective dose of approximately 0.05 mSv per procedure, well below annual occupational limits.
  • The hands receive the highest dose, followed by the eyes and thyroid.
  • Cumulative career dose can become significant for high-volume trauma surgeons performing hundreds of fluoroscopy-guided procedures.
Clinical Implication: Individual procedure doses are low, but cumulative career exposure requires active dose management through ALARA principles.
Limitation: Dose monitoring compliance was variable; actual cumulative doses may be higher than measured.
Source: Giordano BD et al. JBJS Am 2011;93(1):55-63

Pulsed vs Continuous Fluoroscopy Dose Reduction

Randomised Comparative Study
Mahesh M β€’ Radiographics (2001)
Key Findings:
  • Pulsed fluoroscopy at 4 pulses/second reduced dose by 75-90% compared to continuous fluoroscopy.
  • Image quality was clinically acceptable for orthopaedic applications at 4-8 pulses/second.
  • Motion blur was more pronounced at very low pulse rates (1-2 pulses/s) but acceptable for static fracture assessment.
Clinical Implication: Pulsed fluoroscopy should be the DEFAULT mode for all orthopaedic procedures. Continuous mode is rarely necessary.
Limitation: Very low pulse rates may miss dynamic events; higher rates needed during guidewire advancement.
Source: Mahesh M. Radiographics 2001;21(4):995-1012

Dose reduction strategies are well-supported by evidence.

Tip-Apex Distance and Fluoroscopic Assessment

Prospective Study
Baumgaertner MR, Curtin SL, Lindskog DM, Keggi JM β€’ Journal of Orthopaedic Trauma (1995)
Key Findings:
  • A tip-apex distance (TAD) greater than 25mm was associated with significantly higher cut-out rates for sliding hip screws.
  • TAD was measured as the sum of the distances from the screw tip to the apex of the femoral head on AP and lateral fluoroscopic views.
  • Achieving TAD less than 25mm reduced cut-out to less than 2% in the study population.
Clinical Implication: Intraoperative fluoroscopy must be used to measure TAD and confirm screw position β€” the most validated predictor of hip screw cut-out.
Limitation: TAD is one factor; screw position within the femoral head (centre-centre) and fracture reduction quality also matter.
Source: Baumgaertner MR et al. J Orthop Trauma 1995;9(5):401-7

Radiation Exposure: Mini C-arm vs Standard C-arm

Comparative Study
Badman BL, Rill B, Butkovich B, Arreola M, Griend RA β€’ Journal of Orthopaedic Trauma (2005)
Key Findings:
  • Mini C-arm produced approximately 10-fold less radiation exposure than standard C-arm for distal extremity procedures.
  • Image quality was equivalent for hand, wrist, and foot procedures.
  • Patient and surgeon doses were both significantly reduced with the mini C-arm.
Clinical Implication: Mini C-arm should be used for all distal extremity procedures where its field of view is adequate β€” the radiation reduction is substantial.
Limitation: Mini C-arm has limited field of view and penetrating power, making it unsuitable for proximal extremity and axial procedures.
Source: Badman BL et al. J Orthop Trauma 2005;19(10):688-91

Lead Apron Effectiveness for Fluoroscopy Protection

Review
Christodoulou EG, Goodsitt MM, Larson SC, Darner KL, Satti J, Chan HP β€’ Medical Physics (2003)
Key Findings:
  • 0.5mm Pb equivalent lead aprons attenuated approximately 90-95% of scattered radiation at orthopaedic energies (60-120kV).
  • Thyroid shields reduced thyroid dose by approximately 90%.
  • Lead glasses reduced lens dose by approximately 85-90% when properly fitted.
Clinical Implication: Lead aprons, thyroid shields, and protective eyewear are highly effective and should be worn by ALL personnel in the fluoroscopy field.
Limitation: Lead aprons add weight and can cause musculoskeletal problems; lightweight alternatives (barium-based) are available.
Source: Christodoulou EG et al. Med Phys 2003;30(6):1052-64

Clinical evidence supports specific fluoroscopy applications and protective equipment.

Australian Context

In Australia, radiation safety for fluoroscopy is regulated by ARPANSA (Australian Radiation Protection and Nuclear Safety Agency) and state radiation safety authorities. Orthopaedic surgeons using fluoroscopy must hold a radiation use licence (RUL) or work under the supervision of a licence holder, depending on the state or territory.

Australian occupational dose limits for radiation workers are: 20 mSv per year averaged over 5 consecutive years, with no single year exceeding 50 mSv. The lens of the eye limit is 20 mSv per year averaged over 5 years. Personal radiation monitoring (dosimetry badges) is mandatory for all staff regularly exposed to fluoroscopy.

RANZCR and the Australian Orthopaedic Association support the ALARA principle and recommend pulsed fluoroscopy as the default mode for clinical practice. Australian operating theatre standards require appropriate lead shielding availability, radiation warning signage, and regular equipment safety testing by qualified medical physicists.

Exam Viva Scenarios

Practice these scenarios to excel in your viva examination

VIVA SCENARIOStandard

EXAMINER

"An examiner asks: 'Where should you stand relative to the C-arm during fluoroscopy and why?'"

EXCEPTIONAL ANSWER
The surgeon should stand on the IMAGE RECEPTOR (detector) side of the C-arm, not on the X-ray tube side. The physics behind this is that scatter radiation β€” the primary source of occupational dose to the operating team β€” is produced when the primary X-ray beam enters the patient. The scatter intensity is greatest at the beam entry point. In the standard C-arm orientation, the X-ray tube is below the operating table and the image receptor is above. The primary beam travels upward from the tube, through the table, through the patient, to the receptor above. Scatter radiation is produced predominantly at the entry surface of the patient β€” the side closest to the X-ray tube. Therefore, scatter intensity is highest on the tube side (below the table) and lowest on the receptor side (above the table). By standing on the receptor side, the surgeon is on the side with the least scatter. Additionally, the inverse square law means that even small increases in distance from the scatter source significantly reduce dose. Practical implications: (1) When the C-arm is used in the lateral position (horizontal beam), the tube should face away from the surgeon β€” scatter will be directed away. (2) During AP screening with the tube below, the scatter field is directed downward and toward the tube side. (3) Never put your hands in the primary beam β€” and if the beam must pass near your hands, use the lowest possible screening time.
KEY POINTS TO SCORE
Stand on the IMAGE RECEPTOR (detector) side β€” AWAY from the X-ray tube
Scatter is greatest at the beam entry surface (tube side of the patient)
Standard orientation: tube below table, receptor above
For lateral C-arm views: tube faces AWAY from the surgeon
Inverse square law: even small distance increases reduce dose significantly
COMMON TRAPS
βœ—Standing on the tube side (maximum scatter exposure)
βœ—Not knowing that scatter is greatest on the beam ENTRY side
βœ—Not understanding the standard C-arm orientation (tube below, receptor above)
βœ—Putting hands in the primary beam
VIVA SCENARIOStandard

EXAMINER

"You are performing an intramedullary nail for a tibial shaft fracture and need to place the distal interlocking screws using fluoroscopy."

EXCEPTIONAL ANSWER
The freehand technique for distal interlocking screw placement in intramedullary nailing uses the 'perfect circle' principle. First, I remove the external targeting jig (which is often inaccurate for distal locking due to nail deformation during insertion). I position the C-arm perpendicular to the limb and rotate/tilt it until the distal locking hole of the nail appears as a perfect circle on the fluoroscopic image. A perfect circle means the X-ray beam is aligned exactly with the axis of the hole. Once the circle is achieved, I mark the skin entry point by placing a metallic instrument (such as a drill) on the skin surface and centering it over the circle on the fluoroscopic image. I then make a small stab incision and advance the drill bit toward the centre of the circle on the screen. If the drill is perfectly aligned with the hole axis, it will appear as a dot centered in the circle. If the drill point moves off-centre, I adjust the trajectory. Once aligned, I drill through both cortices and the hole. I confirm with a lateral fluoroscopic view that the screw crosses the nail hole correctly. Key technical points: (1) Use the lowest pulse rate possible during positioning β€” only brief exposures are needed. (2) Last image hold eliminates repeated exposures during drill adjustment. (3) The hands should be out of the primary beam whenever possible β€” use angled instruments. (4) Once the circle is achieved, remember the gantry position for the contralateral hole.
KEY POINTS TO SCORE
Perfect circle technique: rotate C-arm until the nail hole appears as a circle
Circle means beam is aligned with hole axis β€” drill toward center of circle
Drill appears as a dot when perfectly aligned with the hole
Confirm with lateral view after placement
Last image hold and pulsed mode minimise radiation during the technique
COMMON TRAPS
βœ—Relying on the external jig for distal locking (often inaccurate)
βœ—Not achieving a true circle (oval means the C-arm is not aligned)
βœ—Excessive fluoroscopy time during drill adjustment
βœ—Hands in the primary beam during screening
VIVA SCENARIOChallenging

EXAMINER

"An examiner asks you to discuss the physiological effects of ionising radiation and the annual dose limits for radiation workers."

EXCEPTIONAL ANSWER
Ionising radiation causes biological damage through two mechanisms: direct and indirect effects on DNA. Direct effects occur when radiation directly damages the DNA double helix, causing strand breaks. Indirect effects (more common at diagnostic energies) occur when radiation ionises water molecules, producing free radicals (hydroxyl radicals) that then damage DNA. The biological effects are classified as: (1) Deterministic (non-stochastic) effects: These have a threshold dose below which they do not occur, and severity increases with dose above the threshold. Examples include radiation skin burns (threshold approximately 2 Gy for erythema), cataracts (threshold approximately 0.5 Gy for detectable lens changes over a lifetime), and radiation sickness. These are relevant for interventional procedures with very high fluoroscopy times. (2) Stochastic effects: These have NO threshold dose β€” any radiation exposure carries some probability of effect, and the probability (not severity) increases with dose. The major stochastic effect is cancer induction. The linear no-threshold (LNT) model assumes that even very small doses carry some cancer risk, proportional to the dose. This model underlies the ALARA principle. Australian occupational dose limits for radiation workers: effective dose of 20 mSv per year averaged over 5 consecutive years, with no single year exceeding 50 mSv. Lens of the eye: 20 mSv per year (recently reduced from 150 mSv). Skin: 500 mSv per year. Extremities: 500 mSv per year. The general public limit is 1 mSv per year. These limits are set conservatively below levels at which deterministic effects occur, while maintaining stochastic risks at acceptable levels.
KEY POINTS TO SCORE
Two mechanisms: direct DNA damage and indirect damage via free radicals from water ionisation
Deterministic effects have a threshold (burns, cataracts, radiation sickness)
Stochastic effects have no threshold (cancer) β€” LNT model
Australian occupational limit: 20 mSv/year averaged over 5 years
Eye lens limit: 20 mSv/year (recently decreased from 150 mSv)
COMMON TRAPS
βœ—Not differentiating between deterministic and stochastic effects
βœ—Stating that there is a safe threshold for cancer risk (LNT model says no threshold)
βœ—Not knowing the Australian/international occupational dose limits
βœ—Not mentioning the recent reduction in the lens dose limit

Fluoroscopy Principles β€” Exam Day Reference

High-Yield Exam Summary

C-arm Basics

  • β€’X-ray tube below, image receptor (detector) above = standard orientation
  • β€’Stand on the IMAGE RECEPTOR side (away from the tube)
  • β€’Flat-panel detector (FPD) is superior to image intensifier (II) β€” less dose, no distortion
  • β€’Mini C-arm: approximately 1/10th dose of standard C-arm for extremity procedures

Dose Reduction (CLAPTD)

  • β€’Collimate β€” narrow beam to region of interest
  • β€’Last image hold β€” review stored frames, not new exposures
  • β€’ALARA β€” use minimum technique factors for adequate image quality
  • β€’Pulsed β€” 50-90% dose reduction vs continuous
  • β€’Time β€” brief exposures, release pedal between checks
  • β€’Distance β€” inverse square law: 2x distance = 1/4 dose

Radiation Safety

  • β€’Lead apron: 0.5mm Pb equivalent attenuates 90-95% of scatter
  • β€’Thyroid shield: 90% thyroid dose reduction β€” mandatory for all
  • β€’Lead glasses: 85-90% lens dose reduction
  • β€’Personal dosimetry badge is mandatory for radiation workers

Dose Limits (Australian)

  • β€’Occupational: 20 mSv/year averaged over 5 years (max 50 mSv any single year)
  • β€’Eye lens: 20 mSv/year (recently reduced from 150 mSv)
  • β€’Skin: 500 mSv/year
  • β€’General public: 1 mSv/year

Dose Increases With

  • β€’Magnification mode (2-4x increase β€” use sparingly)
  • β€’Continuous vs pulsed fluoroscopy
  • β€’Greater patient thickness (ABC compensates with higher output)
  • β€’Closer distance to the X-ray source
Quick Stats
Reading Time67 min
Related Topics

Plain Radiography Principles

Radiation Safety in Orthopaedics

CT Imaging Principles

Imaging the Knee β€” Systematic Approach